[교정장비] Water Triple point cell
Water Triple point cell은 가장 중요한 고정점이며 ITS-90 및 열역학 온도 범위에 공통적인 유일한 기준점입니다. 기준점은 모든 온도 실험실에 필수적인 참조점입니다. Jarrett-Isotech의 Cell이가장 좋은 표준이며, 저급한 장치의 교정에는 적합하지 않습니다.
세부사양
Premium Water Triple Point Cells - Borosilicate Glass | These Premium Cells can be calibrated to an uncertainty of +/- 0.000070°C (0.07mK) A11-50-270*, A13-50-270*, B11-50-270, B11-65-270*, B13-65-270, B16-65-270 *also available in Quartz Glass. |
Reproductilibity | The equilibrium temperature of a cell will repeat to within ±0.000,02°C of the mean equilibrium temperature. |
Stability | After equilibrium is reached, the mean temperature of the inner melt of an ice mantle will remain constant to within ±0.000,01°C for as long as the mantle can be preserved (up to 90 days in some instances). |
Further Laboratory Water Triple Point Cells | These Cells have shorter immersion depth and can be calibrated to an uncertainty of ±0.000100°C (0.1mK). B12-40-210, B12-46-210. |
Accuracy | The equilibrium of temperature of the Jarrett-Isotech Triple Point of Water Cell is within ±20μK of the 2005 definition as described in CCT/05-07/rev(2). Isotopic Analysis can be provided to give the exact composition within ±3μK. |